The auto-calibration systems on cheaper consumer sensors also cause issues if they rarely see air that has low CO2 levels. Because these sensors can't measure absolute CO2 levels, only relative CO2 levels, they provide an absolute figure by looking for the lowest CO2 concentration they've seen over a period of time, usually around a 72 hour rolling window. This works acceptably if the sensor is frequently exposed to…
Auto calibration can be disabled, popular sensors that can do this are MH-Z19 and MH-Z14 (also cheap ones).
My understanding is that with single-NDIR sensors (like the MH-Z14 and 19), the auto-calibration is intended to overcome gradual particle buildup and beam degradation in the sensor chamber. While disabling it will prevent the scenario I described, you'll instead end up with gradual sensor shift as the sensor ages. I guess this could be minimised by manually calibrating the sensor outdoors on a regular basis.
Dual-NDIR sensors split a single beam into two chambers, so any degradation of the sensor beam affects both measurements, and the particle buildup in both chambers should also be approximately equal over time, so they should remain accurate over an extended period without any requirement for calibration. I built mine about 4 years ago and I do occasionally check to make sure they read ~400ppm when placed outdoors, last check was around 420ppm which suggests they're behaving reasonably well as they age.