>I worry it's more than a medical photograph as it's continuous high sample rate video, but I'm not an expert. Would be curious to know. Typical fluoro skin entrance exposure rate to go through someone's hand is on the order of 1-3 millisieverts per minute. With a more advanced detector (like this one) it may be lower, but increasing the frame rate or resolution will tend to require higher dose rate. The associated r…
I've wondered how much the average X-ray dosage dropped since they stopped using film and switched to detector-based imaging. Anyone know?
NCRP 172 (published 2015) provides a direct comparison of digital to film for some examinations. For upper GI fluoroscopy without oral contrast the DRL for film is 3.9 mSv and for digital is 1.5 mSv. I think this is roughly typical.